Dear Wolfram, dear Heiko,


Here are some comments from me about the "13C working standards".

The oxygen in the elemental analysis is needed for several functions. On 
the one hand for the oxidation of the tin capsule and this oxidation 
causes the temperature increase inside the tin capsule, so with the 
sample. Second, to quantitatively react the sample to the analysis gas 
(e.g., N2, CO2, SO2). During the combustion or decomposition process, 
inorganic samples sometimes overoxidize, e.g. SO3 in the reaction. This 
is then reduced in the same reactor or a second separate by means of 
elemental copper to the analysis gas. There must always be enough O2 
available for oxidation and quantitative burning. If you want to further 
increase the temperature inside the capsule, you can use V2O5 as an 
additive in the sample or similar.This also applies to free excess 
oxygen. However, it also consumes the copper reactor.

It is always important to complete the conversion to the sample gas. 
This can be checked by weight and area and element content of the substance.
If it is less than 100%, one should check for by-products at the exit of 
the oven.
Thus, organic and inorganic substances in the elementary analyzer should 
be measurable on isotope ratios and are.

Best,
Matthias

Am 07.12.2018 um 15:05 schrieb Wolfram Meier-Augenstein (pals):
>
> Dear Heiko,
>
> An interesting observation and worth looking into further, especially 
> since it suggests excess O2 to increase temperature in this instance 
> outweighs fundamental principles of chemical reaction equilibrium.
>
> The conversion of carbonate to CO2 can be written as 2 CO3 à2 CO2 + 
> O2.  Since O2 is a product of this conversion, increasing the amount 
> of O2 present in the system and thus on the product side, would shift 
> the equilibrium of the reaction to the side of the educt resulting in 
> an incomplete conversion and thus isotopic fractionation.  This is 
> exactly what one observes when converting nitrate into NOx in an EA in 
> the presence of O2, which has been described by Gentile and Siegwolf.
>
> However, be that as it may, my main concern with using inorganic salts 
> (minerals) in an EA-IRMS to scale anchor organic compounds is that it 
> does not meet with the Identical Treatment principle.  (A) Obviously, 
> inorganic compounds are not a match for organic compounds be it from a 
> temperature stability or be it from an elemental composition (make-up) 
> point of view. (B) Because of that, one way or another matrix effects 
> may impact on results.
>
> There are plenty of organic reference materials (RMs) available that 
> can be used to scale anchor d13C values of organics measured by EA-IRMS.
>
> It is my personal conviction, the use of inorganic RMs such as 
> carbonates should be confined to scale anchor results of stable 
> isotope analysis of carbonates by acid digest. Analysis of CO2 
> released by acid digest of these materials is after all how the d13C 
> (and d18O) values of these inorganic RMs were determined in the first 
> place.   Speaking of acid digest, CO2 released from carbonate RMs can 
> of course be used as reference gas when analysing off-line combusted 
> organics in Dual-Inlet instruments. ;-)
>
> Best,
>
> Wolfram
>
> *From:*Stable Isotope Geochemistry [mailto:[log in to unmask]] 
> *On Behalf Of *Heiko Moossen
> *Sent:* 07 December 2018 13:35
> *To:* [log in to unmask]
> *Subject:* Re: [ISOGEOCHEM] preparation of working standards for d13C
>
> Dear Wolfram and others,
>
> I know that EA-IRMS analyses of NO3 salts often yields better results 
> if the oxygen pulse is turned off. Wolfram cites Gentile et al., 
> (2013) on this subject in his book. However, when CaCO3 is analysed 
> (e.g. NBS19 or its replacement IAEA-603) at least in my lab we keep 
> the oxygen pulse on. In fact we also add WO3 (some labs add (V2O3) to 
> the CaCO3 to provide the reaction with further oxygen. During the 
> oxygen pulse the tin cup compusts in a flash and that further 
> increases the temperature to well above 1020 degrees. I seem to 
> remember reading that during the flash temperatures of around 1800 
> degrees are reached for a very short time, and I think that reaching 
> these temperatures is important to convert the CaCO3 quantitatively to 
> CO2 gas. When we have problems with our EA, e.g. the timing of the 
> oxygen pulse is off, we find that our d13C measurements of CaCO3 are 
> also off. Therefore I would argue that the oxygen pulse during EA-IRMS 
> analyses of carbonates is important.
>
> In this regard I am currently looking for publications that have 
> described the flash procedure, and the temperatures that are reached 
> during the flash. Maybe someone can point me towards publications if 
> they exists. In the German language literature I have found 
> "Glühfarben" (glow colour) tables that link the colour of steel in 
> steel producing furnaces to its temperature. Does something similar 
> exist for other metals (e.g. Tin for Tin capsules)?
>
> Have a nice weekend,
>
> Heiko
>
> Dr. Heiko Moossen, /Dipl.-Chem./
>
> Stable Isotope Laboratory
>
> Max-Planck-Institute for Biogeochemistry
>
> Hans-Knoell-Str. 10
>
> 07745 Jena
>
> Germany
>
> Tel: +49-3641-576400
>
> Fax: +49-3641-577400
>
> P.O.Box 100164, 07701 Jena, Germany
>
> ------------------------------------------------------------------------
>
> *From:*Stable Isotope Geochemistry <[log in to unmask] 
> <mailto:[log in to unmask]>> on behalf of Wolfram 
> Meier-Augenstein (pals) <[log in to unmask] 
> <mailto:[log in to unmask]>>
> *Sent:* 28 November 2018 08:39
> *To:* [log in to unmask] <mailto:[log in to unmask]>
> *Subject:* Re: [ISOGEOCHEM] preparation of working standards for d13C
>
> Dear Mattheus,
>
>
> The d13C values of the RMs described in Arndt's 2016 paper were scale 
> normalized to the VPDB scale as defined by NBS19 and LSVEC. This does 
> not necessarily mean they were analysed against these RMs although one 
> participating lab did so using optimised EA conditions. Other 
> participating labs used RMs such as USGS40 and USGS41 as scale anchors 
> for scale normalization to VPDB. Internationally accepted RMs such as 
> USGS40, IAEA-CH-6 or IAEA-CH-7 are all anchored on the VPDB/LSVEC scale.
>
> In my book I advised not to use inorganic RMs to scale anchor organic 
> samples as this does not meet with the Identical Treatment principle 
> (due to potential matrix effects).
>
> I also pointed out what the issues are and how to overcome these.
>
> Carbonates do not "burn differently". They do not burn at all. 
> Carbonates represent the highest state of oxidation carbon can assume. 
> It is therefore impossible to burn = oxidise carbonates any further.  
> What happens in an EA is a thermolytic breakdown of carbonate into 
> carbon dioxide and oxygen. For this to happen in a way to yield "true" 
> d13C values the reactor temperature has to be above 1000C and the 
> oxygen pulse must be switched off. While some EA-IRMS systems can 
> apply individualised run parameters for each samples, others may not 
> offer this option (e.g. O2 pulse active for one sample but not active 
> for the next).
>
> Using organic RMs for organic samples avoids these issues.
>
> BTW, Arndt and I have recently published a paper on the subject of 
> arMs and their use in Isotopes Environ. Health Stud.
>
> https://www.tandfonline.com/doi/full/10.1080/10256016.2018.1538137
>
>
> Hope this helps.
>
>
> Best,
>
> Wolfram
>
>
> ******************************************************
> Prof. Dr W Meier-Augenstein, CChem, FRSC
> Stable Isotope Forensics & Analytical Sciences
>
> Robert Gordon University
> School of Pharmacy and Life Sciences
> The Ian Wood Building
> Garthdee Road
> Aberdeen
> AB10 7GJ
>
> E-mail: [log in to unmask] <mailto:[log in to unmask]>
> ________________________________
> From: Stable Isotope Geochemistry <[log in to unmask] 
> <mailto:[log in to unmask]>> on behalf of Matheus Carvalho 
> <[log in to unmask] <mailto:[log in to unmask]>>
> Sent: 28 November 2018 05:23:43
> To: [log in to unmask] <mailto:[log in to unmask]>
> Subject: [ISOGEOCHEM] preparation of working standards for d13C
>
> Dear all, I am a little confused about the use of international 
> reference materials for the preparation of working standards. I read 
> the paper by Schimmelmann et al (2016) Anal. Chem. 88, 4294. If I 
> understood it right, they measured the new standards (all organic 
> substances, like caffeine and glycine) in comparison with 
> international standards like LSVEC and NBS19 for d13C. But I also read 
> in the book Stable Isotope Forensics by our estimated contributor 
> Wolfram Meier-Augenstein (excellent book, by the way) that we should 
> not use these materials for organic sample analyzes because they burn 
> differently in the EA reactor, this being the very reason to use the 
> new standards in the Schimmelman paper. But if the standards in that 
> paper were determined against the inorganic substances, why should I 
> trust them more than an organic standard that I determine myself using 
> LSVEC and NBS19?
> Thank you,
>
>
> Matheus C. Carvalho
> Centre for Coastal Biogeochemistry Research, Southern Cross University
> Lismore, NSW, Australia
> Editor for Plos ONE, Heliyon (Elsevier),HardwareX (Elsevier)
>
> [https://docs.google.com/uc?export=download&id=1gZPj7RqUCWOBRFU33c-cVzD20lPvExtN&revid=0B5B3M6j5V1pWVjBYdnNLMlhGR3labkRMdFVBdXhvdkwzZ3lJPQ]
> Osmar, the open source microsyringe autosampler: 
> bit.ly/2QVnDJU<https://t.co/RmmDBIanx4>
>
> Book: Practical Laboratory Automation Made Easy with AutoIt
> http://www.wiley-vch.de/publish/en/books/ISBN978-3-527-34158-0/
> [https://docs.google.com/uc?export=download&id=0B5B3M6j5V1pWN3EwODJjTjVTVUk&revid=0B5B3M6j5V1pWZmxoQzhEVVJuMEpVYUR5c2N3elBadGx5aVVnPQ]
>
>
>
>
> ________________________________
>
> This email has been scanned for spam and viruses by Proofpoint 
> Essentials. Click 
> here<https://eu1.proofpointessentials.com/index01.php?mod_id=11&mod_option=logitem&mail_id=1543382698-OIp9VeIw6wNy&r_address=w.meier-augenstein%40rgu.ac.uk&report=1> 
> to report this email as spam.
>
>
> ________________________________
>
> Robert Gordon University has been awarded a TEF Gold award for the 
> quality of its undergraduate teaching and learning, placing it in the 
> top 20% of Universities in the UK
>
>
> Robert Gordon University, a Scottish charity registered under charity 
> number SC 013781.
>
> This e-mail and any attachment is for authorised use by the intended 
> recipient(s) only. It may contain proprietary material, confidential 
> information and/or be subject to legal privilege. It should not be 
> copied, disclosed to, retained or used by, any other party. If you are 
> not an intended recipient then please promptly delete this e-mail and 
> any attachment and all copies and inform the sender. Please note that 
> any views or opinions presented in this email are solely those of the 
> author and do not necessarily represent those of Robert Gordon 
> University. Thank you.
>
> ------------------------------------------------------------------------
>
> This email has been scanned for spam and viruses by Proofpoint 
> Essentials. Click here 
> <https://eu1.proofpointessentials.com/index01.php?mod_id=11&mod_option=logitem&mail_id=1544189780-zQwLI29Wy56q&r_address=w.meier-augenstein%40rgu.ac.uk&report=1> 
> to report this email as spam.
>
-- 
***********************************************
Dr. Matthias Gehre

Head of the Laboratory for Stable Isotopes (LSI)
Department for Isotope Biogeochemistry
Helmholtz-Centre for Environmental Research - UFZ
Permoserstrasse 15
04318 Leipzig
Germany

Fon: ++49(0)341 235 1361
Fax: ++49(0)341 235 45 1361