We have analyzed ~300mg samples in our Costech before. Our EA has a modified 25mm reactor in the top half, which allows for a little more ash buildup before samples get out of the hot spot.

 

With a stock 18mm reactor you likely will get out of the hot zone after just a few samples. I suggest running without an ash insert to get the maximum ash volume possible for your samples. An ash vacuum will allow for longer maintenance intervals.

 

As to your oxygen loop, semi micro may very well be enough O2. It is the C in the sample, not the physical size that seems to be the primary driver of oxygen need. We typically select the O2 loop based on the weight of C in the sample. Some soils could probably combust fine on a micro loop even.

 

It would be a good idea to make sure your EA timing is optimized, sample delay, run time, etc. to get thorough combustion. Also try to target standards, gas pulses all in the low signal range, ideally across a range to create a linearity curve.

 

Best regards,

 

Ben

 

 

Benjamin Harlow

Manager, Stable Isotope Core Laboratory

Washington State University

367 Eastlick | Pullman, WA  99164-4236

Office: 509-335-6161 | Lab: 509-335-6154 | Fax:509-335-3184

http://www.isotopes.wsu.edu

 

 

 

From: Stable Isotope Geochemistry <[log in to unmask]> On Behalf Of Sarah Grosshuesch
Sent: Monday, September 09, 2019 11:29 AM
To: [log in to unmask]
Subject: [ISOGEOCHEM] 300mg sediment EA-IRMS

 

Anybody ever run a 300mg sed sample w/ very low (<0.5%) carbon and nitrogen by EA-IRMS ?

I am wondering if a Costech 4010 ECS semi-micro loop for oxygen will be adequate for 

combustion.

 

Thanks -  Sarah

--

Sarah Grosshuesch
Researcher | Large Lakes Observatory |
[log in to unmask]
2205 E. 5th St., Duluth, MN 55812
University of Minnesota Duluth
[log in to unmask] | 218-726-7206